star probe.
Do you just need to know how to name the stylus, or do you need complete instruction on probe qualification?
Z- (if you had one) would be #1, Y+ stylus would be named #2, X+ would be #3, Y- would be #4, X- would be #5
Go to CMM tab > Stylus system icon > pick up the MasterProbe > Choose the Reference sphere > choose Ref. sphere position (click the rotation and angle that matches your actual sphere). Click Qualify stylus > follow the prompts to qualify your MP. Pick
up the star probe > click Qualify stylus (manually qualify each styli).
Is there is any special in qualifying star probe instead as usual qualification of each tips??.
I have not worked with star probes....
Ive done star probes on an xxt gold head, where each leg of the star was its own probe. Just havent done it on an RDS where i have to tilt the head 90 deg.
OK, I see. In that case I name my stylus (for example) 90.0_45.0_Y+_1.0 That
indicates an A rotation of 90, a B rotation of 45, and the 1.0 stylus facing Y+