Jump to content

star probe.


---
 Share

Recommended Posts

Do you just need to know how to name the stylus, or do you need complete instruction on probe qualification?

Z- (if you had one) would be #1, Y+ stylus would be named #2, X+ would be #3, Y- would be #4, X- would be #5
Link to comment
Share on other sites

Go to CMM tab > Stylus system icon > pick up the MasterProbe > Choose the Reference sphere > choose Ref. sphere position (click the rotation and angle that matches your actual sphere). Click Qualify stylus > follow the prompts to qualify your MP. Pick
up the star probe > click Qualify stylus (manually qualify each styli).
150_1af3016d4f73bdf0142a3688b55a1a3e.png
Link to comment
Share on other sites

Please sign in to view this quote.

Is there is any special in qualifying star probe instead as usual qualification of each tips??.

I have not worked with star probes....
Link to comment
Share on other sites

Ive done star probes on an xxt gold head, where each leg of the star was its own probe. Just havent done it on an RDS where i have to tilt the head 90 deg.