Good Morning,

We have a long probe that is 1mmX120mm. We calibrate it using a small reference sphere, tensor mode with a force of 30mN and 20% dynamic, but starts to bend when touching off on the sphere.
Does anyone know if that is normal or if I need to change the force and the dynamic percentage? Looks like its going to snap.

Thank You,
Nixon Uruchima
Link to comment
https://qualityforum.zeiss.com/topic/15410-1mmx120mm-probe/
Share on other sites

Please sign in to view this quote.

We've gotten it qualified before but it does flex/bend when doing so. Just wanted to see if that was normal or what would the normal procedure be.
It is something Zeiss facility used for one of our parts when we had them build us a program.

The lowest that ive done is 30mN. anything lower and the head starts going backwards after a few points.
I just worry this thing will snap at any time from all the flexing.
Link to comment
https://qualityforum.zeiss.com/topic/15410-1mmx120mm-probe/#findComment-83124
Share on other sites

Please sign in to view this quote.

3033_d081f294944dfc4b71a2b4f48de09414.png
When it touches off on the sphere it touches 2 times on the same spot before moving to the next spot.
The first time it backs off right after touching it but the second time it puts more force I guess until it bends and then relaxes back into position.

I usually get a deviation of .0018 every time I calibrate this thing.
Link to comment
https://qualityforum.zeiss.com/topic/15410-1mmx120mm-probe/#findComment-83136
Share on other sites

Please sign in to view this quote.

Yes, the bending is normal. Watch the lower right-hand corner of the LCD screen on the joystick controller box. You'll see the probing force values used for each point taken. The variation in the point readings help determine the bending map of the stylus.

Thanks for the picture. Do you have a backup probe handy? 😃
Link to comment
https://qualityforum.zeiss.com/topic/15410-1mmx120mm-probe/#findComment-83142
Share on other sites

Richards idea is actually better until you bring the PO for the FixAssist and the 30mmxØ20mm Rotary Joint to management.

If you really want to be accurate without the equipment then you can make cylinder on the reference sphere with the shank of the stylus to get the exact A1/A2 angles.
Link to comment
https://qualityforum.zeiss.com/topic/15410-1mmx120mm-probe/#findComment-83160
Share on other sites

I'd definitely want to verify the stylus accuracy by checking a ring gauge (ID) or plug gauge (OD) around the same size you're using it to check to see if it's actually measuring the size correctly.

I've attached an older document on how to use the gauge correction feature but, it should still be relevant.
Could probably select the F1 key and search for the current instructions.

Gauge Correction Qualification inst.doc