probe qualification/calibration


Recommended Posts

I know there is a way to do it, just need some direction from the experts here! How do I on a daily basis qualify my probes for the individual positions that have been set for each stylus system with out having to select and manually probe thier start postion in relation to the cal sphere? Is there a program for this or otherwise?? Thanks in advance for the help!
Link to comment
https://qualityforum.zeiss.com/topic/16886-probe-qualificationcalibration/
Share on other sites

Aslo, if you just need a single probe that is currently qualified done real quick, pick it up from the rack and press the icon of a set of calipers in the top right hand corner of the Probe Qualification Page.

Pressing this will prompt you to re-qualify the current probe in all five positions.