probe qualification/calibration


---
 Share

Recommended Posts

I know there is a way to do it, just need some direction from the experts here! How do I on a daily basis qualify my probes for the individual positions that have been set for each stylus system with out having to select and manually probe thier start postion in relation to the cal sphere? Is there a program for this or otherwise?? Thanks in advance for the help!
Link to comment
Share on other sites

Thanks so much! I have no idea why none of this was covered in the basic training course from Zeiss, maybe the figured I was smart enough to figure it out LOL
Link to comment
Share on other sites

Aslo, if you just need a single probe that is currently qualified done real quick, pick it up from the rack and press the icon of a set of calipers in the top right hand corner of the Probe Qualification Page.

Pressing this will prompt you to re-qualify the current probe in all five positions.

download.jpg