machine calibration delay
If you have a known calibrated artifact you can make some checks.
But even calibrated ring gages etc.. will only tell you so much. I have master parts that I run before and after a calibration. I keep them locked in a cabinet with the reports.
If you have something like that. IF not maybe make one up when you get it calibrated and verified..
My 0.02
Investing in a quality artifact (purchased or self-made) is the best way to go, if calibrations are going to be delayed by months.
One can make the argument that bi-monthly reference checks can be used to justify extending the calibration cycle. Essentially, the artifact will "pay for itself", if you are able to transition from a 12 month cycle to an 18 month cycle.
As other's here have posted, use a part with known results.
The one we have here is a master part we keep locked up and run if there is a question on the accuracy of our CMM.
It has 115 characteristics uses 5 different stylus and about 30 different articulations.
If results match we feel our machine is good and this method passed muster on our last audit.
Greg K