I don't have an accurate answer for that,
We were developing scannings since ever with active sensors (Prismos in Lab and Gagemax in shopfloor), no significant problems between machines.
Then because advantages in cost and space, I installed in 2014 couple Duramax in shop floor using a program developed in a Prismo with VAST sensor..

Was my first experience with Vast XXT and the results were not the expected, after some time we achieve acceptable results without the need to use different programs.

Not sure exactly how much this contributes but at that time we noticed a better performance using sensitive calibration, and reading the Calypso documentation we found that was recommendable for high precision, slow speed and small stylus diameters; In that case we were evaluating some distances (+/- 0.015 mm) extracted from curves.
Link to comment
https://qualityforum.zeiss.com/topic/17254-calibration/#findComment-96640
Share on other sites

In the case of DuraMax and O-Inspect machines, I'm not as certain on the sensitive vs standard being the main issue. Most of it stems from the probing dynamics. The mechanical machines tend to need to be ran in to 20-30% range.

If you are working with probes smaller than 0.5mm then 10-15% is recomennded.