Bad scan??


---
 Share

Recommended Posts

Below is a snip of a scanned ID. I can't explain what's going on with the points gathered. Everything looks normal during the run itself. The probe isn't touching anything but the (clean) ID, but that's not what the points show. I use this probe for several other features, and there are no problems with anything else. There are a series of ID's all at a similar height, and they all look like this...

Thoughts/ideas?

Dirty scales? Faulty probe head?

Scan.JPG

Link to comment
Share on other sites

I've rerun the ID with a couple of different probes. Very similar results. It's not shanking out. The ID pictured is a through hole, there's nothing to hit.
Link to comment
Share on other sites

Passive? I'm not sure what you're asking, or how to answer you. It's a VAST XXT TL3. The head is 2-3 years old. I've never seen scans like this, not even close. I've propped the fixture up about 2 inches, and reran the ID's and got similar readings....
Link to comment
Share on other sites

Possibly a broken "collision protection plate" inside the XXT. I have seen normal scans show this, due to a broken plate interfering with the movement of the probe. Simply removing the XXT and shaking it slightly should tell you if there is one "swimming around" in there!
Link to comment
Share on other sites

There are 10 diameters, of various sizes and heights, but all oriented left to right.

They all look similar.

I feel like an idiot child having to ask, but how hard do I shake it? I shut it all down, took the XXT off, and shook it a bit harder than I'd shake a bottle of OJ. I didn't heard anything.
Link to comment
Share on other sites

Lol. No worries, I doubt anyone has ever told you to do this! I have found that these plates can get stuck in the spring mechanism, so a little shake may not produce a rattle if that is the case. Otherwise, a slight shake should produce a rattle, which is the tell tale sign. If still unsure, you can actually see these plates if you look inside the end of the XXT.
Link to comment
Share on other sites

Please sign in to view this quote.

Move the sensor by placing it near your ear and if you hear something inside the sensor then very surely you will be damaged by an insurance or retainer of the restraint system, this type of sensors already in two occasions have given me that problem of damaging that suspension system by say it in a way
Link to comment
Share on other sites

Is it possible that you are out of travel ? For example, if a part is not in the nominal position, the defined path cannot maintain contact with part that is grossly out of position.
Link to comment
Share on other sites

I'm with Chad. Try and replicate the measurement scenario with a ring gage.
If it's showing bad that way, you've eliminated the part as a variable.
Link to comment
Share on other sites

I check a ring gage, with the same probe, and in the same orientation as the feature, yesterday. The ring looked gorgeous. Another detail is that the "bad" spots are always at the top of the scans. I can rotate the part and/or the probe, and the "bad" spots are always at the top.

Also, today, the machine is fumbling probe changes. Smaller (lighter) probes aren't staying in the probe rack, the magnetic head is pulling them forward, out of the rack when it drives away.

A coworker (unhelpfully) suggested the CMM could be on a ley line...
Link to comment
Share on other sites

That fact that the gage ring measured good would indicate to me that it is either the part or the measurement strategy.

The fact that the probe head is acting funny would indicate that it is the probe head or machine.

Something is missing because these facts contradict each other.
Link to comment
Share on other sites

It's extremely frustrating Chad.

I'm also trying to work on several other things, so this problem doesn't really have my full attention, which it SHOULD have, if I have a hardware problem. I'll tweak the program, and see how single points, and partial scans look....
Link to comment
Share on other sites

Hello
if you have done the verification on a ring and the defect is not present you don't have hardware problems, if re-doing the measurements with the piece (and the probe) rotated on another axis and the defect follows the rotation then the problem is the piece.
Link to comment
Share on other sites

Has the CMM been crashed? Or depending on your set up, one of the length or segmented measurement systems been compromised? Sometimes the segmented measurement systems are strips that are easily accessible. And little things such as cleaning, etc can even compromise them. Leaving you still able to calibrate and run parts correctly, but those areas compromised could cause trouble in various ways depending on how bad it is.
Link to comment
Share on other sites