Quick question about calibration programs. We have always located the sphere with the Master Probe then run each probe individually through the full qualification (12 positions using RDSCAA). However, I have no problem creating the calibration program and adding the "probing system qualification" and qualify passive. If you qualify passive, it only does the 6 positions not 12. Is this accurate enough or should we continue to use full qualification on each probe individually? Would appreciate any suggestions!

Thank You!
Link to comment
Share on other sites

Test the re-qualification of an RDS-CAA stylus system by reporting the Z value of a point on top of your part at A0/B0. Then, copy and paste this point and edit to A90/B0, and another at A180/B0 and another at A-90/B0 and so on. Compare the reported Z values. Then, do a Full Qualification and run your program again.

Please post the variation.
Link to comment
Share on other sites

Your re-qual looks better than the full qualification. In years past, the re-qual was considerably worse. Maybe the issue was fixed. Granted this is only in 1 direction involving only the A rotation but I've seen as much as .002" deviation on my machine, as well as a couple of our customers. Once, I tried to incorporate many other probe angles but never got around to it.
.