Duramax XXT TL3 Styli limits


Recommended Posts

I don't use our Duramax's much but, get called when they're having issues.

Currently, they're trying to use a probe that is well within the 15g weight limits but two has 100mm carbon fiber probes hanging laterally to each side on a 50mm down ext. (see attached picture).

The probe qualifies and runs just fine but, keeps getting the ole "not correct amount of pressure not detected" error where it repeats measuring a bore ID multiple times over and over.

I've read that the TL3 is not supposed to have a probe exceeding 65mm laterally so, that's most likely why it's not working for them.

Tried qualifing the probe with all 3 different pressure settings to no avail, probe still does same thing and STD less than 0.0003mm

Does anybody have any work arounds where this could be made to work?111_febd6f89e6bf1293b64fb1b76a1f3889.jpg
Link to comment
https://qualityforum.zeiss.com/topic/21180-duramax-xxt-tl3-styli-limits/
Share on other sites

Robust setting in the features setting editor,...Hmmmmm. I'll give that a try and thank you.
I've always used an active scanning head, didn't even know there was a "probing behavior for passive scanning" option in there. Is isn't on my active scanning computers but, it's there on the Duramaxipad computer.
×
×