Two different values when ran in AutoRun and normal Calypso.


Recommended Posts

This doesn't make any sense at all.

One program, it is setup in AutoRun.

It has some PCM to determine if the part is tilted in the X axis.

measure("YZ Finder")
TRYAGAIN:
measure("Tilt Finder A")
measure("Tilt Finder B")
if abs(getActual("Tilt Finder A").x-getActual("Tilt Finder B").x)>0.004
message("Please Re-Position part.  Ensure that the Flange of the part is fully to the left of the V Block and X Bore is straight down")
goto TRYAGAIN
endif

When ran in Autorun it reports .00432xxxxx. (Gets stuck in an endless loop due to false value).

When run in Calypso it reports .00234xxxxx. (Works as written).

Part is mounted in the fixture and secured, and I can get the same results at will.

Same program, same points, the program doesn't change, only the result of the actual probed points.

Tilt Finder A and Tilt Finder B.

Part is cylindrical and its axis is in X. Plane surface is ZX.

Tilt Finder A touches a plane surface at Y0, X0 and Z+.945

Tilt Finder B touches a plane surface at Y0, X0 and Z-.945

How is this possible?

 

Please sign in to view this quote.

No, I have not, not sure what that will do but I will try it.

This issue is on both CMM's that utilize Autorun for families of parts, I suspect it has been an issue for many years, and no one noticed.

  • 2 months later...

I don't think I understood your question initially.

The two points are not patterns, but I do use a lot of patterns for other features.

The points touch a plane, one at +Z and the other at -Z, this is to detect if the part is tilted in the ZX axis.

The points are after the Base Alignment.

×
×