Jump to content

Calypso 8.0: Planner - Simulation


---
 Share

Recommended Posts

Just an FYI - Planner/Simulation will not longer simulate correctly if the model for your part is still set to machine coordinate zero (top-left corner). In the past, this was never a requirement, but it appears that it is forced in 8.0 now. 

You can move this at any point, it will not have any effect on your program if you do it before or after it is programmed. 

Before model is moved (hard to see, but essentially it just doesn't move). 

2026-02-1008-56-32.thumb.gif.9537d4d090af5c74889a32d6d7390811.gif

 

After model is moved. 

2026-02-1008-57-46.thumb.gif.e55eb9a632b68978c742d2ace3e6e11c.gif

 

In the Measurement Plan Simulation - select Modify CAD Model - on the Positioning tab - select your CAD Model - enter the appropriate Translation values (you can always start with half of your measuring volume) - click Apply. 

image.thumb.png.76b7025a113ac2bfa56cb6a9123f9cc3.png

  • Thank you! 2
Link to comment
Share on other sites

Richard,

Thank you for your research and the short videos.
 

Fortunately, we have not had to simulate any machines so far, so the component was and is always at the machine origin. The displacement has rarely worked properly for us. In most cases, the coordinate system was displaced and this displacement could not be reversed.

However, the "old" Planner version did not correctly simulate exceeding the measuring space (end stop), nor did it detect collisions with its own geometry (table, side parts, probe rack). Our CONTURA 700x700x1000 was not available as a simulation model. I don't know whether this variant is included in CALYPSO 8.0. But if the planner no longer works without the correct CMM model, there's no point in installing it in the first place.
It is not always necessary to load and simulate large graphics. I believe it is imperative to switch off the machine simulation.

Link to comment
Share on other sites

Karsten, 

The machine simulation is turned on by default, but regardless you have to perform the translation with or without it on. I only showed the machine in this case to showcase why it wasn't working correctly. 

Link to comment
Share on other sites

 Share

×
×