TP of hole Pattern
I would probably measure as cylinders and intersect with the OD. If you measure as circles, then the "depth" at which it is measured plays a huge role in it's position evaluation. If the part isn't machined really close to perfect, I could see it becoming an issue.
I'd also consider changing the coordinates to polar.
What Calypso version are you on? I'm not sure how the best fit handles a pattern where the axis of the features are not parallel/perpendicular to the primary datum.
  • Like! 1
https://qualityforum.zeiss.com/topic/26298-tp-of-hole-pattern/#findComment-148736
You cannot use the old GD&T Bore Pattern Best Fit for holes that are perpendicular to the axis of rotation - it will only truly work (with a pattern at least) if they are parallel to the axis of rotation.
As long as you don't need any MMB/LMB evaluation - I would first use Geometry Best-Fit to the DRF (allow rotation), then evaluate the Position of each cylinder back to the Geometry Best-Fit - you will need to either use the Polar evaluation or the Special Rotation to align the axis for each hole.
  • Like! 1
https://qualityforum.zeiss.com/topic/26298-tp-of-hole-pattern/#findComment-148892