Howdy, I have a 425mm long probe mounted horizontally. Ive been getting some weird repeatability issues on this bore and im looking for some advice on how to make this program more repeatable. 

The length of the measured bore is around 337mm long and we have to use a line scan for the given geometry. You can see the crest at the end of the bore indicated by a 20X mag of points and the cylindricity plot.
Every time i run the program the crest is either in a different location radially or not there at all. Any ideas of what could be causing this and how to avoid it.

Scanning direction is pulling (is it more stable to scan in a push direction?) at 7mm/s 
DK10 L425 
Contura 
Vast XTR. 

3.jpg

2.jpg

1.jpg

Link to comment
https://qualityforum.zeiss.com/topic/27059-long-probe-repeatability-issues/
Share on other sites

---

Please sign in to view this quote.

I suppose I could use single points but were trying to keep cycle times down and we feel like scanning is faster. What would you recommend for a point count if we moved to single points. 

Im worried about increasing the magnitude of error on the measurement by going in through both sides. Historically, in our experience measuring from 2 sides has lead to more discrepancies in our results than good.  

---

Please sign in to view this quote.

I did not know this! Thanks for the nugget of wisdom. 

Rigidity is at 1.263, so pretty poor... by probing dynamics is at 30%, measuring force is at 200mN, Qualifying in Tensor. After playing around with it this gave me the best sigma value. But im open to try new qualification settings if you've had better luck. 

---

Please sign in to view this quote.

I remember one experience trying to qualify a flimsy star-type stylus system on an XXT. I couldn't get a good qualification and assumed a sensitive probing dynamic would improve my results, but it wasn't until I tried a robust setting that I get a better qualification.  It didn't make sense at first but then I thought maybe more pressure helps stabilize the probe.  I don't know if that works across the board, but it worked in that instance.

---

I have a similar setup, except its 345mm long.

I use:

M5, Styli straight, silicon nitride sphere, tungsten carbide shaft

Stylus straight M5, DK14 L65

626115-1401-065
and a:

Extension Ø11, ZEISS REACH CFX® 3

Extension M5-DG11-L280-1x-M5

626107-1280-100
 
This also uses a 40mm extension and a 50gr counterweight on the opposite end.
I am scanning 13 cylinders in specific locations (They are broken by grooves). This creates a critical Datum D.
I will have to look at my rigidity, but I am not having any issues.
 

Please sign in to view this quote.

 

×
×