Reference Sphere Info Not Updated


Recommended Posts

---

Please sign in to view this quote.

The Update/Do not update stylus data at definition of sphere position means that when referencing the sphere’s position with the MasterProbe the Geometry of the             MasterProbe will be updated based off this sphere. An instance where you may want to set this to Do not update would be if you are adding a sphere for an optical             sensor. You don’t want the MasterProbe’s geometry to change based off this optical sphere instead you want the MasterProbe to tell the CMM the location of the             sphere thus the optical sensor can be qualified but we want the MasterProbe to remain the same. 

  • Like! 1
---

The way it was explained to me was that your main reference sphere should be set to "yes" and any additional sphere should be to "no". For example I have an additional 8mm sphere to calibrate small probes and on that one I have it set to "no". Is this correct? I remember I was trying to find an answer even on the old forums and it was never really clear to me.

---

I was told, and it should still be on the forum (many years ago maybe 2015 or so), you turn off Stylus Update when you are re-qualifying a stylus.

You have to reference the Master Sphere, but you do not change the Master Probe data to keep all of your styli matched as if they were all qualified together.

---

I have never changed any of this setting on any of the machines under my control.  The above issue is one someone else's machine.  I am looking into what that setting might be and will update the post. 

 

The 2nd part of my curiosity is does this setting have any affect on qualifying the other stylus systems?

×
×