Jump to content

DRAFT ANGLE


---
 Share

Recommended Posts

You can either scan it like a cone or like a line.

For cone you then report half of it's angle.
For line you then report angle between features ( old Calypso has that close to profile callouts )

  • Like! 1
Link to comment
Share on other sites

Looks like the angle is relativ to the the extremly short area (1,924). That is in general not a good idea because minor deviations (form) on that short area will result in big deviations in the angle.

Theoretically the angle should be the same as to the area "11,14". But you sould have of course detailled knowledge about the usage/installation to just replace the the one reference of the angle.

I would measure both and compare to get a feeling for the situation.

Link to comment
Share on other sites

Please sign in to view this username.

There is a runout callout to A, then i assume that 10° is half of that cone or angle of line to A axis. I can not tell more since i don't see if it's full circle/cone or just a half.

Link to comment
Share on other sites

From what I see you have about 34.576 of length to scan; I don't see the angle as only in the 1.924 reference dim.

I cone with 3 or 4 circular scans should be sufficient.

Link to comment
Share on other sites

Please sign in to view this quote.

You cannot, pratically. Circle surface scanning is too litle. You have to start form a solid calypso allignment starting from Datum A, then scan the cone on two far section, then on charatheristic choose "semi angolo del cono".

Link to comment
Share on other sites

 Share

×
×