Recommended Posts

We just got the ATOS Q 8M with ScanPort, and I am trying to calibrate the MV50 lens. I can get Position 1 to pass. My question is: how do I move the ScanPort for the next position 2 without exiting the calibration window?"

MV50 Calibration.jpeg

MV50 Calibration_Position 1.png

Link to comment
Share on other sites

The ScanPort has its own app for calibration. It will handle it for you. Though I'm not certain if the normal ScanPort app handles the MV50 by default. 

Link to comment
Share on other sites

MV50 will probably be problematic due to vibrations and movement checks.

You can use the scanport app --> full calibration.

Or use a tritop.

 

image.thumb.png.a31ad4c6c9de9b5ce3fe2bd7e88fdf5f.png

Edited
Link to comment
Share on other sites

  • 2 weeks later...

Jochen, thank you for the suggestion. I am using MV270 lenses, Calibration worked until I had to tilt 45 deg, and I can't figure out why it's saying 'WRONG TILT DIRECTION' and how to proceed. I attached a image with the calibration panel set-up.

IMG_3437.jpeg

Screenshot 2026-08-21 153018.png

Link to comment
Share on other sites

Please sign in to view this username.

 i think he is allright.

That 4 points in the middle is oriented correctly ( at least from lower camera images ) and accordingly to picture hints.

What i don't understand is why there are 4 camera screens displaying different things.

Link to comment