Jump to content

Air filtering systems


---
 Share

Recommended Posts

We have oil escaping our dryer/filter systems and bleeding into our Contura (size - 12/24/8. serial # 732207560438) and Accura (size - 16/30/10. serial # 201209810136).  We want to set up something to capture the oil for both CMM's. what is the best option for this? Is it best to have one system in place for both CMM's?

Link to comment
Share on other sites

We have 4 stages before Accura own filters ( i hope i got that right )

1. 25um ( C238-F11/3 )
2. 5um ( C238-F12/3 )
3. 0,1um ( MX2-F10 )
4. 0,1um carbon ( MX2-F11 )

  • Like! 1
Link to comment
Share on other sites

---

After encountering the same problem, we also purchased air dryers to go with our CMM devices.

We have a single system for both devices, an annual maintenance agreement, and the filters are changed regularly.

 

 

.image.png.f163411eac933ef56c7734bb23fdc29a.png

Link to comment
Share on other sites

  • 1 month later...

The most oil i see in 2nd tube and 1st filter is oily after usage ( when changing )

3rd and 4th is not visibly used, but air is better for machine. We are having service in 2y periods - not sure if they are changing machine's filters ( we have some in drawer )

Link to comment
Share on other sites

 Share

×
×
  • Create New...