[Ro...] Posted Friday at 10:10 PM Share Posted Friday at 10:10 PM Has anyone encountered this error when trying to do a full calibration with the ScanPort? I get this same error when it's on position 12. On the right side it says that it needs to decrease distance, but this is ScanPort automatic calibration, I don't have the option to move the sensor closer. Any guidance is appreciated. Link to comment Share on other sites More sharing options...
[Ri...] Posted Saturday at 02:58 PM Share Posted Saturday at 02:58 PM Please sign in to view this username. Is this a GOM Scan 1 or an ATOS Q? I ran into issues with the last GOM Scan 1 that I was using where the arm ran out of room to move closer. There was nothing I could do, so I ended up calibrating with the tripod. I believe there is a different mounting bracket for the GOM Scan 1 to the ATOSQ, so make sure you’re using the correct one. Link to comment Share on other sites More sharing options...
[Jo...] Posted 2 hours ago Share Posted 2 hours ago You could move the plate ... Which sensor and MV are we talking about? Link to comment Share on other sites More sharing options...
[Im...] Posted 1 hour ago Share Posted 1 hour ago Hi Roy , try moving the first starting position of the calibration plate up by about 1 cm so that there’s still space to move at position 12, then it should work. A similar problem occurred for us when calibrating the GOM Scan 1, of course make sure to check if you have the correct adapter installed (different one for the Atos Q and a different one for the GOM Scan 1). Link to comment Share on other sites More sharing options...