[Jo...] Posted September 8 Share Posted September 8 In training they told me probes less than 1mm should qualify on an 8mm sphere, and larger probes on a 30mm sphere. Does anyone have any wisdom as to why? What exactly are the bad effects of qualifying on too large/small of a sphere? We are troubleshooting some wierd problems and realized we have been qualifying a 2.5mm probe on an 8mm reference sphere for a little while. Sigma is always low. Wondering how far I need to run this up the NCR tree..... Link to comment https://qualityforum.zeiss.com/topic/27384-qualifying-on-too-largesmall-of-a-reference-sphere/ Share on other sites More sharing options...
[Ma...] Posted September 8 Share Posted September 8 I think mainly due short probe shaft. Link to comment https://qualityforum.zeiss.com/topic/27384-qualifying-on-too-largesmall-of-a-reference-sphere/#findComment-155384 Share on other sites More sharing options...
[DW...] Posted September 8 Share Posted September 8 Please sign in to view this quote. Please sign in to view this username. I would imagine qualifying 180° of a small stylus on a large sphere would lead to shaft shank out problems more times than not. Link to comment https://qualityforum.zeiss.com/topic/27384-qualifying-on-too-largesmall-of-a-reference-sphere/#findComment-155386 Share on other sites More sharing options...
[Ch...] Posted September 9 Share Posted September 9 I agree with the posts above when using a small stylus (sub 1mm) on a large 30mm ref sphere. A small stylus on a large ref sphere runs the risk of shanking out if the shaft is stepped. Example image below. If you can ensure this isn't occurring on your setup then I wouldn't think there should be any other limitations. I also don't think you should see any issues when using a large stylus (over 1mm) on a small ref sphere (8mm). Unless it is so large that it's contacting the shaft of the ref sphere itself when scanning from the side. Maybe others have experienced differently but I have used upwards of 5mm diameter styli on my 8mm sphere without issue at the 180/0 position. 2 Link to comment https://qualityforum.zeiss.com/topic/27384-qualifying-on-too-largesmall-of-a-reference-sphere/#findComment-155408 Share on other sites More sharing options...
[Ri...] Posted September 24 Share Posted September 24 Use a step stylus. I use 20mm OAL 10mm L and 5 ML Qualify full 180° all the time on a 30mm Sphere on two Micuras. Link to comment https://qualityforum.zeiss.com/topic/27384-qualifying-on-too-largesmall-of-a-reference-sphere/#findComment-156073 Share on other sites More sharing options...
[Jo...] Posted September 25 Author Share Posted September 25 We *are* seeing significantly smaller sigma values when qualifying on the smaller sphere, but both values are well under an acceptable value for our measurements. If this matters, I have not been able to see it metrologically yet. Link to comment https://qualityforum.zeiss.com/topic/27384-qualifying-on-too-largesmall-of-a-reference-sphere/#findComment-156143 Share on other sites More sharing options...