evaluating a 45° angle between planes.
We're having some trouble agreeing on an angle result.
Depending on what the other plane is, the CMM measures the angle between 43° - 47°
image.png.cb9fc7f44544a5261c84fe9619177a3b.pngimage.png.7c0c1734025bd2a937965695d2d5b559.png
I'm told by the engineer that it's actually a near-perfect 45°.
The customer uses a 3d scanner and their results are very close to my cmm results.
Is there a super secret way to evaluate angles differently?
Here's a visualization of the chamfer's profile.  I added random datums to constrain the DoF.
image.png.69a926dfc35b58525ba9a79f05948556.pngimage.png.29ba28eda70fd1116b082cd25b2350a0.png
If it's just a chamfer, then i would stick to the closest feature to this chamfer. In other words it's local angle
If part is thin it can bend by fixture force or machined features can bend via machining forces. Then chamfer is to none machining force.
So if it's chamfer of a hole - you can compare angle to plane at hole or hole itself.
For me this is too small to accuratelly measure.
For 0,5mm chamfer i do only one point in the middle of chamfer.
Make intersection from faces to obtain sharp corner.
Place a touch point and direction set to 45°
From that point place theoretical 3d line and intersect from both sides. Now you can measure distance from both directions.
For our usage is this fine.
  • Like! 2
I wouldn't try to do a profile evaluation of it... The first feature would be the top face not any symmetry or anything like that... I would either scan a line on top face and a line on the chamfer itself with 0.010" speed and 0.0175-.020" length... Then just do "Angle Between Features" characteristic...... Have the filtering on and evals be LSQ for both features.....
I've tried line to line angle as well, and get nearly the same thing as the plane to plane angle.
I'm having a hard time understanding the cad evaluation.  Wouldn't the shape of the angle measurement be the same as angularity?  I'm also not sure why the cad eval doesn't change color when the characteristic is out of tolerance.
I'm doing grid scans on the planes.
Reason I go with line to line is to avoid as much of form error as possible.... Unless features are large enough faces to be considered planes.
If you need to continue scanning the planes, then I would make sure that your planes filter out the form errors properly as well as use LSQ eval instead of Outer Tangential.
For CAD Eval: Once you set it click on a different characteristic and then back on the one you wanted to look at. Also make sure that you have "color gradient" turned on.
Angularity itself is checking the form and orientation of the face compared to the nominal angle. It is no different than checking parallelism (form/orientation at 0°) and perpendicularity (form/orientation at 90°), everything in between those two angles falls under angularity.
If the callout is a simple chamfer check, then doing grid scans and checking the form/profile is probably overdoing it.... I would try to picture it as if you set the part on the comparator how would you check/verify your result and then have CMM mimic that.
Hi
.  It looks like you've really been diligent on analyzing this problem.  I like the form plots.
Regarding your two last questions:
1. "Wouldn't the shape of the angle measurement be the same as angularity?"
That's an understandable assumption.  However, no, the shape of the scanned deviations for an angular measurement are fundamentally different from those in a GD&T angularity characteristic.
The reason for this relates to the shape of the tolerance zones.  Angularity establishes a linear tolerance zone that is controlled by a basic angle driven by the datums.  In the case of your chamfer feature, the shape of the tolerance zone is two parallel planes. The .001064 deviation is the quantity of linear deviation from the center of that coplanar tolerance zone.
Regarding colors: The magnitude (farther or closer from nominal) of deviations in the angularity characteristic will assign the color gradient based on departure from the coplanar tolerance zone.
2. "I'm also not sure why the cad eval doesn't change color when the characteristic is out of tolerance."
The way that the color gradient in Calypso angular characteristic should be assigned is in an arc-shaped path relating to the angle.  However, for angle measurement, all we get is a form plot that shows us the form deviation of the two features of interest.  This is why there is no color gradient assigned, and it's all green.  This CAD evaluation is still useful; however, because it shows you how the planes are calculated according to their form error, and significant form error can be a signal for out-of-tolerance results.
I hope that answers those two questions.  Chamfers are important features that are often downplayed as "simple."  Also, the evaluation algorithm that is assigned to each of the features that construct your chamfer is very important.  Outer tangential method will be prone to rocking the planes and introduce high variation in results.  If you're using high density scanning, which appears to be the case, LSQ may give you more stable results.  Either method may be valid, but it helps to be aware of the effects.
Chamfers are also a rare case where I sometimes enable the "constraints" options in the evaluation tab of the features to prompt the characteristic to only consider deviation along vectors that matter functionally.
I'm interested to hear how this chamfer measurement problem develops for you.
Thank you Jeff and Marcel!  I'm very grateful for both of your responses.
I completely agree that I'm over doing it.  Typically I would also just make line scans, since it's just an angle and length measurement, nothing to it.  I've been attempting to convince my co-workers that no matter where I take the line scan, it's not 45°.
Constraining both planes in the characteristic evaluation was my issue.  It's outputting a result that the engineer can agree with.
Like I said, my go-to scanning methods usually work.  I've never had problems measuring angles between planes before, it's cone angles I have issues with evaluating.  I was getting the same thing on the comparator and keyence too.
The customer is still 3d scanning the feature and will probably tell us it's out.  But I did what I could.
Thanks again for all the information!
I'm lost again attempting more research.
The results are in tolerance if the angled feature is constrained, but the cad eval looks warped.
The results are out of tolerance if neither are constrained, but the cad eval looks fine.
For Cone angle I would capture the cone with either line scans or helix path with start/stop about 20-30% from the edges. (this is to avoid surface finish from affecting angle readings)
Second thing automatically go and constrain normal vector for the cone in the feature evaluation and then just report either cone angle (included) or half cone angle
If you constrain the chamfer plane (it constrains it normal to the alignment it was captured in to say it's perfect to that alignment) then you are essentially checking your top plane to the alignment of the part.
The cad eval issue might be the magnification factor way too high (or possibly set to auto)
Put the part on a comparator.
If its what engineering states, 45°, you are wrong
If its nearly your results, you are correct.
End of story.
Your first two graphs (bar graphs) tells me there are issues with the middle plane and the plane in the lower bar graph image that could contribute to the results you get off the CMM.
I have, and match the CMM and Keyence results.  And the customer's results.  Our engineers think the planes need to be constrained.  Even though I've never had to constrain a plane before.
I'm scanning and evaluating this thing muuuuultiple ways.  The results are always similar.
Part of me want's to believe the engineer because the cad eval shows the points within their zones.  But constraining the features, and the result is conforming.  But the eval looks non-conforming.  However earlier this week, Jeff informed me about angles showing form instead of result.  That would mean the form is good and the measurement is accurate?  And Marcel was saying 'when constraining the angled features, the form is compared to the alignment.'  So the constrained view is what the part looks like?  buuuut measures the angle in tolerance.
non-constrained:
constrained:
That "constrained" picture seems like wrong nominal vector - that second picture is not showing correct 45°
I would believe that "non-constrained" is constrained. How you got to show this image. By "Form" or profile?
They're just angle characteristics.  The cad eval is enabled with tolerance Windows.
The angle is oot.  So the top image results are correct.  I was expecting the top image to look like the bottom image.
How big is magnification? 200? I rather use cylinders than those points.
So all features are having correct nominal vector and same feature's alignment, right?