Drives shutting off during movement between measurements


Recommended Posts

I have noticed that if I have the Speed set to 300 in the program, it will occasionally error out, and kick off the drives.  If I set it at 200 or 160, its fine.   I have also noticed this happening during probe qualifications (List Qualification).  If I turn the speed control knob to 75%, it runs fine.  Once i go back to 100%, it will alarm out every now and then. 

 

I have cleaned all the ways with lint free cloth and 99% isopropyl alcohol, and the clothes are coming off fairly clean. ( i do this every other day).

Is there a global setting I can change to keep the maximum travel speed to 250 or less?

Edited

As far as I know, the cut off pressure is 5 bar for Contura. So even if it drops to 4.99 or below momentarily, the drives will turn off. We had similar issue where it would shut off at lower speeds even. Zeiss tech told me the same thing as Luis and I have not had any issues after the adjustment.

Apart from air pressure the symptoms point to a lag issue. If the CMM starts moving with high acceleration and the belt (or whatever drives the axis) slips slightly, you get a difference between the distance the axis should have travelled and how far it actually moved according to the scales. This difference is constantly monitored by the controller, and if it exceeds a certain amount, the drives are shut off.

Edited
Posted (edited)

I increased the regulator on the CMM to an air pressure of 5.05 and that seems to have corrected the issue.  I also increased the regulator coming out of the air dryer ( shop air > air/oil separator combo filter > dryer > hard air line > air/oil separator combo > CMM) to 7.0 to ensure I don't lose air pressure when both CMM's are running.