[Ja...] Posted 2 hours ago Share Posted 2 hours ago (edited) I have noticed that if I have the Speed set to 300 in the program, it will occasionally error out, and kick off the drives. If I set it at 200 or 160, its fine. I have also noticed this happening during probe qualifications (List Qualification). If I turn the speed control knob to 75%, it runs fine. Once i go back to 100%, it will alarm out every now and then. I have cleaned all the ways with lint free cloth and 99% isopropyl alcohol, and the clothes are coming off fairly clean. ( i do this every other day). Is there a global setting I can change to keep the maximum travel speed to 250 or less? Edited 2 hours ago Link to comment Share on other sites More sharing options...
[Ze...] Posted 2 hours ago Share Posted 2 hours ago Maybe it's an air pressure issue? Link to comment Share on other sites More sharing options...
[Ja...] Posted 2 hours ago Author Share Posted 2 hours ago I thought that as well, and I forgot to mention that I verified the air pressure only dropped to 4.98 bar when moving. Normally it sits between 5.00 and 5.04 bar. Link to comment Share on other sites More sharing options...
[Lu...] Posted 1 hour ago Share Posted 1 hour ago i had issues with that the zeiss techinician told me to ajust at 5.05 max 5.1 Bars at 5.05 i have no issues in our new brand contura Link to comment Share on other sites More sharing options...