[Ja...] Posted 10 hours ago Share Posted 10 hours ago (edited) I have noticed that if I have the Speed set to 300 in the program, it will occasionally error out, and kick off the drives. If I set it at 200 or 160, its fine. I have also noticed this happening during probe qualifications (List Qualification). If I turn the speed control knob to 75%, it runs fine. Once i go back to 100%, it will alarm out every now and then. I have cleaned all the ways with lint free cloth and 99% isopropyl alcohol, and the clothes are coming off fairly clean. ( i do this every other day). Is there a global setting I can change to keep the maximum travel speed to 250 or less? Edited 10 hours ago Link to comment Share on other sites More sharing options...
[Ze...] Posted 10 hours ago Share Posted 10 hours ago Maybe it's an air pressure issue? Link to comment Share on other sites More sharing options...
[Ja...] Posted 10 hours ago Author Share Posted 10 hours ago I thought that as well, and I forgot to mention that I verified the air pressure only dropped to 4.98 bar when moving. Normally it sits between 5.00 and 5.04 bar. Link to comment Share on other sites More sharing options...
[Lu...] Posted 9 hours ago Share Posted 9 hours ago i had issues with that the zeiss techinician told me to ajust at 5.05 max 5.1 Bars at 5.05 i have no issues in our new brand contura Link to comment Share on other sites More sharing options...
[Ja...] Posted 2 hours ago Share Posted 2 hours ago As far as I know, the cut off pressure is 5 bar for Contura. So even if it drops to 4.99 or below momentarily, the drives will turn off. We had similar issue where it would shut off at lower speeds even. Zeiss tech told me the same thing as Luis and I have not had any issues after the adjustment. Link to comment Share on other sites More sharing options...
Recommended Posts
Please sign in to comment
You will be able to leave a comment after signing in