Hi all,

If I put a Probing System Qualification in my probe calibration program for the Masterprobe at the beginning, and use Geometry Requalification, does it update the reference sphere position and then calibrate the Masterprobe? I'm trying to understand if I need to manually find the RS position each time with the Masterprobe at the beginning of an autocal program, or if it finds it during the geometry requalification? If so, is there a search distance it uses to locate the RS? If not, how do I update the RS position automatically at the beginning of the program before autocalibrating all the other probes?

Hope that made sense. 

Link to comment
Share on other sites

I have one calibration program with a VAST XT Gold sensor; we automatically qualify the MasterProbe.

The Reference Sphere (RS) is always in the same place on this setup (It's on a Micura 5x5x5). The RS is removed when running products and put in place when running qualification every Monday. It is located on an Inspection Arsenal plate.

We do not have to manually reference the RS, it is all automatic.

We use default search distance that is setup when you define the MasterProbe.

Edited
  • Like! 2
Link to comment
Share on other sites

Please sign in to view this quote.

I was suspicious that the magnetic plates could do the job, but we use the magnetic inspection arsenal sphere holder as well and it is great!  You can just snap the sphere on and off the table and it remembers the position. Expensive, but Inspection Arsenal is just fantastic stuff. 

Link to comment
Share on other sites

Let me try and explain a little better what I'm asking. Please correct me if I have anything wrong with my understanding. 

If you remove the reference sphere and then replace it on the table, even if it's at the same location, you need to find/redefine the reference sphere position by loading the MasterProbe and running Ref Sphere Position. This is done with a Manual hit on top of the Reference Sphere. This ensures that points are taken at the correct vector and location on the RS during probe calibration by redefining the RS location. After that, all probe calibrations are done based on that RS position. This RS position is saved until it's redefined by using the same process again. 

I want to automate that process in a probe auto-calibration program. If I put a Probing System Qualification in my program as the first characteristic, choose the MasterProbe as the first probe to check, and select Qualify Passive Stylus, for example, does the software redefine the RS location? What about Geometry Re-Qualification? I know it will initially drive/move to the last known location of the RS, but if it doesn't update it then how would I do that at the beginning of my program? The only options in the Probing System Qualification dialogue for the MasterProbe, or any probe, are Qualify Passive Stylus and Geometry Re-Qualification. 

After perusing the internet and asking AI I'm getting all sorts of conflicting answers. Some say the RS position is updated, some not, some to put a macro in (which I can't find), etc. 

Any insight on how to do this correctly would be much appreciated. I can't find any sort of definitive source/paper on this from Zeiss or any other source. 

Thank you!

Link to comment
Share on other sites

I'd be interested in this as well. It's been my experience that the only time the RS information gets updated is when a manual qualification happens. I can do any of the other tips I have, but when it comes to the RS, only a manual calibration works. Although I can fixture a RS to be in the same place, it's never in the exact same place. It's only close. When checking really tight tolerance stuff, only being close can make the difference between pass and fail. That's why I insist on my operators doing only manual qualifications when it comes to the RS. If there's a better way to do it, I'm all ears.

Link to comment
Share on other sites

I just tested Passive Qualification, and the Reference Sphere's position does update when done manually and automatically.

I tested this on a 543 with the 4mm pallet and removed and installed the reference sphere between runs.

 

RS_old.PNG

RS_after_man.PNG

RS_after_auto.PNG

  • Like! 1
Link to comment
Share on other sites

I can't speak for XXT geometry requalification, but for passive, and also for tensor and 6 points  on an VAST/XT. All of these run fine automatically and update the RS position. We use a setup like this for many years now. I guess any type of (re)calibrating the master probe is synonymous to updating the sphere location.

Link to comment
Share on other sites

Yes it will do the master probe (Sphere Master Probe relationship) and update sphere position in the calibration program.

I have setup 3 spheres in a calibration program works great. 