[Be...] Posted November 5 Share Posted November 5 I recently had a problem that took me way too long to figure out. I had a part that measured great except for the few features that were measured with the angled probe. I beat my head against it and eventually discovered that the issue was the calibration with the regular straight 3mm probe on a 75 mm stem. It was off by almost 0.03". I think that what had happened was that somebody had not revalued the reference sphere before calibration. The interesting thing is that the numbers on the calibration were great. It looked like a good calibration. Good sigma etc. With roughly 100 features measured using over 10 probe orientations of basically every approach angle with that probe it agreed with itself. Holes on one side of the part were directly in line with holes on the other side of the part. I would expect that to not be the case. The Y direction was off by almost 0.03" (not to mention the issues in X though Z was good). So when I am probing from the +X side vs probing from the -X side with a generic probe, if the reference sphere location was wrong by 0.03" I would expect the holes on the two sides to be off by 0.06" relative to eachother. I think the CAA just does a ton of magic to bring those into alignment so that anything with that probe is basically good to other things with that same probe. But if you use another probe with a regular calibration ... especially a good calibration ... the two probes will disagree by the offset in the reference sphere location, or about 0.03" in this case. Does that sound right? Does anybody else have more insight or experience with this? I cannot afford the cmm time to truly test the theory. Link to comment https://qualityforum.zeiss.com/topic/27828-caa-probe-calibration-magic/ Share on other sites More sharing options...
[Mi...] Posted Friday at 06:35 PM Share Posted Friday at 06:35 PM I ran into this recently when tracking down an error and spent along time trying to wrap my head around it. I'll leave it to more experienced RDS users (former Vast exclusive guy) to confirm but yes. When I ran my verification program from standard calibration styli against RDS-CAA Styli I saw a major move but in general the CAA calibrations were just not as good because of how it's making a mathematical model essentially to try and cover all of the positions. I'm looking at transferring what I can to a list calibration as everything here previously was set up to CAA except for 1 stylus. Still figuring that out 1 Link to comment https://qualityforum.zeiss.com/topic/27828-caa-probe-calibration-magic/#findComment-158500 Share on other sites More sharing options...
[Ze...] Posted Friday at 08:55 PM Share Posted Friday at 08:55 PM If you're seeing .03" difference, even using CAA, you have another problem. The difference between using CAA compared to individual probe angles not using CAA is nowhere near that amount of difference. This sounds like a calibration issue. Do you use more than 1 reference sphere? What is your calibration procedure? Link to comment https://qualityforum.zeiss.com/topic/27828-caa-probe-calibration-magic/#findComment-158504 Share on other sites More sharing options...
[Ja...] Posted Monday at 03:40 PM Share Posted Monday at 03:40 PM It's been an awfully long time since I've been on an RDS machine, but I do remember that it was not at all uncommon to see 0.005" error using CAA which is why I quit using it. I would agree with Zen Cat that if you're seeing that much error, there's something else going on. Link to comment https://qualityforum.zeiss.com/topic/27828-caa-probe-calibration-magic/#findComment-158526 Share on other sites More sharing options...
[Be...] Posted Monday at 05:31 PM Author Share Posted Monday at 05:31 PM Please sign in to view this quote. The issue was that they never valued the reference sphere when they did the CAA calibration. Then they go back and calibrate the other probe, which is only used at two angles but they value the reference sphere first. So the two probe calibrations don't agree with each other at all. I was beating my head against the issue, and it didn't make a lot of sense to me. So I decided to map the error a little more carefully and just manually measured the reference sphere with both probes from various angles. The CAA probe agreed with itself in positions 1, 2, 3, and 4. The odd probe agreed with itself in three angles that I tested. But the two probes disagreed by about 0.03". So I brought in the master probe and checked the sphere. That agreed with the odd probe. So I recalibrated the CAA probe and retested and the discrepancy disappeared. The thing that really bothered me about it is that the numbers on the CAA calibration looked really good. Sphere diameter and sigma were both great. The offsets (XYZ) looked really funky and I had trouble making that out. After some work, I came to understand those numbers, and I would be able to better understand any problems that can be diagnosed by looking at those in the future. 1 Link to comment https://qualityforum.zeiss.com/topic/27828-caa-probe-calibration-magic/#findComment-158532 Share on other sites More sharing options...
