Help, I deleted the probe!


Recommended Posts

So I needed to delete one probe angle off the list and accidently deleted the whole thing.  I have recreated the probe.  I recreated all the individual angles and such, but I am going back through and calibrating them.  It is taking too long.  I only know how to calibrate angles one at a time.  I usually use CAA probing in the area I work in and was helping out and ... messed up big time.  

I need to be able to program a CNC calibration for a probe that has never been calibrated at that angle.  I can't seem to figure out how to do that.  I am not certain I could even if the probe was just needing a recalibration.  

Any shortcuts?

Link to comment
https://qualityforum.zeiss.com/topic/27872-help-i-deleted-the-probe/
Share on other sites

Please sign in to view this quote.

list qualification only works with one reference sphere location / orientation.  If the list requires multiple reference sphere locations then you cannot us this option.  Also, it only runs on a new new probe.  If you calibrate 5 or so of the angles, then it is no longer an option.

If anybody is interested, the solution that Zeiss recommended as the fix to avoid future issues was to regularly back up the configuration folder.  We did this before the delete and the process of restoring did not take.  Most of the probes were screwed up when they came back in from restoration and many did not come back at all.  

The list qualification feature only works when a probe is new AND it only works for one location / orientation of one reference sphere.  So for a probe like ours where we have to use four separate reference sphere locations it has very limited use.  Further, because historic calibrations of several probe angles made it through the restore process, the list qualification function is not an option.  

One of the big secrets to fix the problem turned out to be to go to Stylus System Management then settings and fix the reference sphere listed for probes that already had XYZ data.  Once I had that, then all the probe angles that had XYZ data could be calibrated in a program with four functions, one for each necessary reference sphere location.  This meant that it took about 8 hours to run through those calibrations, but I could do others stuff like keep the other CMM running.  

That got a lot of it fixed.  We still had to go through and figure out all the angles that did not get any recovery and add those in one at a time.  Again, list calibration was not an option.  So making the list took almost 8 hours and calibrating all those probe angles about another two hours.

×
×