[Ti...] Posted August 14, 2024 Share Posted August 14, 2024 Good morning everyone, I have a new star probe that the shaft length is only 5mm long so of course I can not calibrate it going to 180 degrees of my sphere, I lowered the sphere coverage but the probe it still trying to calibrate at 180. What am I missing? Link to comment https://qualityforum.zeiss.com/topic/5048-calibrating-a-new-star-probe/ Share on other sites More sharing options...
[Ma...] Posted August 14, 2024 Share Posted August 14, 2024 You can not use dynamic tensor - this will always do 180° coverage Link to comment https://qualityforum.zeiss.com/topic/5048-calibrating-a-new-star-probe/#findComment-16922 Share on other sites More sharing options...
[Ti...] Posted August 14, 2024 Author Share Posted August 14, 2024 Please sign in to view this quote. So where can I change this? Link to comment https://qualityforum.zeiss.com/topic/5048-calibrating-a-new-star-probe/#findComment-16924 Share on other sites More sharing options...
[Ma...] Posted August 14, 2024 Share Posted August 14, 2024 When you are doing manual calibration, then in top middle there is combobox with Dyn. tensor. I think regular "Tensor" will do it. Otherwise there is 6 points. Link to comment https://qualityforum.zeiss.com/topic/5048-calibrating-a-new-star-probe/#findComment-16926 Share on other sites More sharing options...
[Mi...] Posted August 14, 2024 Share Posted August 14, 2024 If this is an XXT machine, you can't, you need a smaller sphere. Link to comment https://qualityforum.zeiss.com/topic/5048-calibrating-a-new-star-probe/#findComment-16928 Share on other sites More sharing options...
[To...] Posted August 14, 2024 Share Posted August 14, 2024 Sounds like you have an XXT sensor, which ALWAYS performs the scanning portion of the Passive qualification at 180°. The angle property only comes in to effect during the 6 points after the scanning portion. However, you've built the bending parameters on the full 180°. As Mike mentioned, a smaller diameter (8mm) reference sphere will be required. Link to comment https://qualityforum.zeiss.com/topic/5048-calibrating-a-new-star-probe/#findComment-16930 Share on other sites More sharing options...
[Ca...] Posted August 20, 2024 Share Posted August 20, 2024 In any case, you will need a small calibration ball with a diameter of 8mm. I had the same problem and was only able to solve it this way Link to comment https://qualityforum.zeiss.com/topic/5048-calibrating-a-new-star-probe/#findComment-16932 Share on other sites More sharing options...