Probe X and Y off location


Recommended Posts

Please sign in to view this quote.

If you run calibration, how much are differences between calibrations? -> possible loose screw with high approach speed
Since reported diameter is "almost" same as it should be, there is no stem touch.

I would try some calibrations to see changes on probe XYZ data. Then lower dynamics to 15-20% and perhaps probing force to 100mN

Please sign in to view this quote.

Have you tried calibrating both probes?
Are you using two different spheres to calibrate both probes?
Is the probe damaged?
Is the 1mm shanking when scanning the ring gage? I know a lot of 1mm probes have a small Measuring Length.
Feels like the calibration or the shanking might be the problem since its off by so much.
×
×