[Ry...] Posted May 30, 2023 Share Posted May 30, 2023 I am measuring the KMG Check artifact and have several features that have failing U-Factors of >1. If you run the KMG artifact at your location and have updated the limit values do you get your program to pass? About my setup: MPE(E0), MPL(R0), MPE(PFTU), MPE(RONt) and MPE(THP) are entered as 0.90+L/500, 0.10, 0.60, 0.60, 0.90 respectively per the calibration cert for the machine. All sizes of artifact features have been entered per the calibration cert for the artifact. I do notice that the report says there is a Measurement Process Factor (MPF) of 1.2, so it is taking the machine limits from above and adding 20% for the artifact check. Currently I have the artifact setup in the X axis orientation. Link to comment Share on other sites More sharing options...