[St...] Posted May 17, 2023 Share Posted May 17, 2023 Looking for some help. I am using an RDS-CAA with VAST XXT, and in one direction the probe is not square to my machine. A+90 seems to be the problem. Any recommendations on how to fix this? Thanks A+90, B-90 = Not square A+90, B+90 = Not square A-90, B+90 = Square A-90, B-90 = SquareCMM Probe.jpg Link to comment Share on other sites More sharing options...
[Da...] Posted May 17, 2023 Share Posted May 17, 2023 How did you square up the 1,2,3 block? Link to comment Share on other sites More sharing options...
[St...] Posted May 17, 2023 Author Share Posted May 17, 2023 Yes, the 1,2,3 block is square to the machine. The first picture was in A +90, B-90. This next picture i am uploading is in A-90, B +90. both probing in the same machine angel, but one is out of square the other is square. Not sure how to fix it.CMM Probe 1.0.jpg Link to comment Share on other sites More sharing options...
[Ri...] Posted May 18, 2023 Share Posted May 18, 2023 Please sign in to view this quote. As long as the 18-pin probe adapter is seated correctly, there isn't anything that you can do short of calling hardware support. Link to comment Share on other sites More sharing options...
[To...] Posted May 19, 2023 Share Posted May 19, 2023 I had something similar to this this happen once when the RDS was replaced. Apparently, the CAA mapping needed to be re-loaded. Zeiss did it remotely. Link to comment Share on other sites More sharing options...