Actually right at this very moment im using a glue gun to "stick" a tall but lopsided part up against some blocks.
I dont use it very often, maybe 1 or 2 times per month, but it always stays close.
This particular part will stand up by itself, but im afraid that probing pressure will cause it to rock, so i just need something to steady the part.
If you know of an aluminum magnet supplier, let me know.

glue1.PNGglue2.PNG

I've used Inspection Arsenal for many many years.
So when I spotted your Master Sphere sitting on an Aluminum plate it reminded me of when I attempted to do that.
What I found, and coincidently a Zeiss Engineer also determined the same as I had.
It was unstable for qualifying probes.
I had a program to automate stylus qualifying, it would take almost two hours to qualify the rack of probes.
After, we started noticing deviations between styli running the same part.
Testing revealed a probability of the aluminum plate expansion/contraction caused deviations during qualification.
Placing the Master Sphere bolted to the granite without the aluminum plate and all the issues went away.
I was able to repeat this issue many times.
A Zeiss engineer discovered this on another of our CMMs while calibrating, we removed the aluminum plate and the issue went away.

contura 4.jpeg

I guess I'm guilty of this myself. I routinely calibrate probes to less than <0.001mm STD and see little to zero variation in an environment kept between 68-69° fahrenheit and (update I just checked it - generally less than 0.002mm in reference sphere position change).
I suppose it would all be tied to how accurate you want your CMM to be correlating to the tightest accuracy required in your facility. Up until about 5 years ago, Zeiss themselves used aluminum probe extensions in great quantities.
The thermal expansion coefficient of aluminum is significant (12.9_10-6 in/(in oF) no doubt but, because most professionally sold CMM fixture plates (R&R, Rayco, ect) are hard-coat anodized, I wonder if this procedure reduces the thermal expansion?

Please sign in to view this quote.

Tom, How tall is the Machine with the lower rack. Management is all bent out of shape because our old Contura G2 work horse is down for calibration. They asked me to get a quote for a new Contura. Problem is the cealing is only 107 inches tall. My guess they would either have to build another space, or raise the roof. This will be interesting. Wish me luck I get another machine.

Please sign in to view this quote.

Not sure what you mean "with the lower rack" The overall height for the the 9/12/8 z is 3060mm (120.5") This is the same for all of the 800mm Z machines. However, Zeiss came out with a 600mm Z version, so I would assume the height is 200mm shorter making it 2860mm (112.6") I don't have the literature for the newer version models. The 7/7/6 and 7/10/6 models are 2658 (104.65")

Please sign in to view this quote.

I meant with the lower table. I accidentally said rack. We have a 10/12/6 Contura G2 that sits on the lower stand that Zeiss used to make. I guess this is not available anymore. We want to keep as close to similar to these dimensions so everything we have running on this CMM could be used on the new machine as well. We are getting quotes for the machine so we will see what happens. Thanks.
  • 7 months later...
Has anyone had trouble switching heads? I am on a Contura with an RDS and we just had a Vast XT installed. I noticed that the CMM gets angry when I switch heads, unless I shut everything down first. Is there a way to change the head without shutting down the whole machine?

Thanks in advance!

Please sign in to view this quote.

.

I have friends down the road from me who are using a new Brown & Sharpe Global that supposedly switches sensors in CNC mode mid-measurement without needing recalibration of probes. This claim sounds audacious to me, and I wonder how accurate it is. If true, it would certainly be a leg up on multi-sensor inspection.

.
You should not have to power the machine down in my experience yet I am not sure if I have conducted this on a Contura 6206 specifically. Accuras, Prismos and Centermax only. Turn off the drives, swap the sensors, turn on drives and change machine tab connection. No need to power off as far as I know. I will investigate and if I am incorrect I will relay that information.